How much patient effort is needed to trigger a ventilator breath when there is 8 cm H2O of unintended positive end-expiratory pressure (auto-PEEP) and a pressure trigger setting of 2 cm H2O?

Prepare for the Mechanical Vent Test with our study tools, featuring multiple choice questions, explanations, and practice exercises. Get ready to ace your exam!

Multiple Choice

How much patient effort is needed to trigger a ventilator breath when there is 8 cm H2O of unintended positive end-expiratory pressure (auto-PEEP) and a pressure trigger setting of 2 cm H2O?

Explanation:
Auto-PEEP raises the baseline airway pressure, so a patient using a pressure trigger must generate enough effort to overcome that intrinsic pressure and still reach the trigger threshold. The trigger will fire when airway pressure falls by the set trigger amount below this elevated baseline. Here, the intrinsic PEEP is 8 cm H2O and the trigger sensitivity is 2 cm H2O, so the patient must produce 8 + 2 = 10 cm H2O of inspiratory pressure to trigger a breath. That’s why the required effort is 10 cm H2O. Without auto-PEEP, only 2 cm H2O would be needed.

Auto-PEEP raises the baseline airway pressure, so a patient using a pressure trigger must generate enough effort to overcome that intrinsic pressure and still reach the trigger threshold. The trigger will fire when airway pressure falls by the set trigger amount below this elevated baseline. Here, the intrinsic PEEP is 8 cm H2O and the trigger sensitivity is 2 cm H2O, so the patient must produce 8 + 2 = 10 cm H2O of inspiratory pressure to trigger a breath. That’s why the required effort is 10 cm H2O. Without auto-PEEP, only 2 cm H2O would be needed.

Subscribe

Get the latest from Passetra

You can unsubscribe at any time. Read our privacy policy